View Full Version : Which pH meter
bo00_ger
April 28th 07, 07:43 PM
thinking about buying a pH meter for continuous operation in a sump. Which
one to get? Pinpoint? Milwaukee?
Wayne Sallee
April 30th 07, 06:33 PM
Pinpoint.
Wayne Sallee
Wayne's Pets
bo00_ger wrote on 4/28/2007 1:43 PM:
> thinking about buying a pH meter for continuous operation in a sump. Which
> one to get? Pinpoint? Milwaukee?
>
>
Pszemol
May 2nd 07, 01:54 PM
"Wayne Sallee" > wrote in message nk.net...
> bo00_ger wrote on 4/28/2007 1:43 PM:
>> thinking about buying a pH meter for continuous operation in a sump.
>> Which one to get? Pinpoint? Milwaukee?
>
> Pinpoint.
Would you care to give any particular *reason* for this choice?
Wayne Sallee
May 2nd 07, 10:23 PM
I don't care for the design of the Miwukee. It's
more toyish in design, and does not have the same
calibration options as a better quality ph meter.
Wayne Sallee
Wayne's Pets
Pszemol wrote on 5/2/2007 7:54 AM:
> "Wayne Sallee" > wrote in message
> nk.net...
>> bo00_ger wrote on 4/28/2007 1:43 PM:
>>> thinking about buying a pH meter for continuous operation in a sump.
>>> Which one to get? Pinpoint? Milwaukee?
>>
>> Pinpoint.
>
> Would you care to give any particular *reason* for this choice?
Wayne Sallee
May 2nd 07, 11:21 PM
A Little More Info About Ph Meters:
Most people calibrate ph meters with two calibration
fluids. I use all 3 (4,7,10). Yes you can calibrate
with just two, but I'm a perfectionist :-)
I relabeled my calibration knobs.
I relabeled the (7) to be
Lower <--> Higher
I relabeled (4,10) to be
4,10 <--> 7
Now here's why:
Imagine you have a yard stick, or a meter stick. Now
imagine that 4 is on the far left, and 10 is on the
far right, and 7 is in the middle. Now with the 7
knob what you are doing is moving that measuring
stick left or right to make 7 line up in the middle
where the 7 is supposed to be. When you calibrate,
you should do the 7 first.
Now imagine that you are stretching and compressing
the measuring stick so that the 7 and 10 line up.
That's what the other knob does. So let's say you
are calibrating with the 7,10 knob, with a 10
solution, and let's say it reads 9.something. And
then you check the 4 with a 4 solution, and it reads
4.0. As you are adjusting with the 10 solution, and
the 4,10 knob, as you move the reading from
9.something, towards the 10, the 4 will be moving to
4.something. So if I have to choose between the
accuracy on the 10 side or the accuracy on the 4
side, then I will chose based on whether I am using
that meter for low Ph fresh water, or for salt
water. A new probe is not likely to be very
inaccurate, but as they age, they start loosing some
accuracy.
Now if you were to purchase a Ph meter that only had
one calibration knob on it, which calibration knob
would you like to be without. I would not want to be
without either one :-)
Wayne Sallee
Wayne's Pets
Wayne Sallee wrote on 5/2/2007 4:23 PM:
> I don't care for the design of the Miwukee. It's more toyish in design,
> and does not have the same calibration options as a better quality ph
> meter.
>
> Wayne Sallee
> Wayne's Pets
>
>
> Pszemol wrote on 5/2/2007 7:54 AM:
>> "Wayne Sallee" > wrote in message
>> nk.net...
>>> bo00_ger wrote on 4/28/2007 1:43 PM:
>>>> thinking about buying a pH meter for continuous operation in a sump.
>>>> Which one to get? Pinpoint? Milwaukee?
>>>
>>> Pinpoint.
>>
>> Would you care to give any particular *reason* for this choice?
Wayne Sallee
May 2nd 07, 11:25 PM
Oops typo, should have been "3.something"
The following is the corrected version :-)
####################
A Little More Info About Ph Meters:
Most people calibrate ph meters with two calibration
fluids. I use all 3 (4,7,10). Yes you can calibrate
with just two, but I'm a perfectionist :-)
I relabeled my calibration knobs.
I relabeled the (7) to be
Lower <--> Higher
I relabeled (4,10) to be
4,10 <--> 7
Now here's why:
Imagine you have a yard stick, or a meter stick. Now
imagine that 4 is on the far left, and 10 is on the
far right, and 7 is in the middle. Now with the 7
knob what you are doing is moving that measuring
stick left or right to make 7 line up in the middle
where the 7 is supposed to be. When you calibrate,
you should do the 7 first.
Now imagine that you are stretching and compressing
the measuring stick so that the 7 and 10 line up.
That's what the other knob does. So let's say you
are calibrating with the 7,10 knob, with a 10
solution, and let's say it reads 9.something. And
then you check the 4 with a 4 solution, and it reads
4.0. As you are adjusting with the 10 solution, and
the 4,10 knob, as you move the reading from
9.something, towards the 10, the 4 will be moving to
3.something. So if I have to choose between the
accuracy on the 10 side or the accuracy on the 4
side, then I will chose based on whether I am using
that meter for low Ph fresh water, or for salt
water. A new probe is not likely to be very
inaccurate, but as they age, they start loosing some
accuracy.
Now if you were to purchase a Ph meter that only had
one calibration knob on it, which calibration knob
would you like to be without. I would not want to be
without either one :-)
Wayne Sallee
Wayne's Pets
Wayne Sallee wrote on 5/2/2007 4:23 PM:
> I don't care for the design of the Miwukee. It's
more toyish in design, and does not have the same
calibration options as a better quality ph meter.
>
> Wayne Sallee
> Wayne's Pets
>
>
> Pszemol wrote on 5/2/2007 7:54 AM:
>> "Wayne Sallee" > wrote in
message
nk.net...
>>> bo00_ger wrote on 4/28/2007 1:43 PM:
>>>> thinking about buying a pH meter for
continuous operation in a sump.
>>>> Which one to get? Pinpoint? Milwaukee?
>>>
>>> Pinpoint.
>>
>> Would you care to give any particular *reason*
for this choice?
Wayne Sallee wrote on 5/2/2007 5:21 PM:
> A Little More Info About Ph Meters:
>
> Most people calibrate ph meters with two calibration fluids. I use all 3
> (4,7,10). Yes you can calibrate with just two, but I'm a perfectionist :-)
>
> I relabeled my calibration knobs.
> I relabeled the (7) to be
> Lower <--> Higher
> I relabeled (4,10) to be
> 4,10 <--> 7
>
> Now here's why:
>
> Imagine you have a yard stick, or a meter stick. Now imagine that 4 is
> on the far left, and 10 is on the far right, and 7 is in the middle. Now
> with the 7 knob what you are doing is moving that measuring stick left
> or right to make 7 line up in the middle where the 7 is supposed to be.
> When you calibrate, you should do the 7 first.
>
> Now imagine that you are stretching and compressing the measuring stick
> so that the 7 and 10 line up. That's what the other knob does. So let's
> say you are calibrating with the 7,10 knob, with a 10 solution, and
> let's say it reads 9.something. And then you check the 4 with a 4
> solution, and it reads 4.0. As you are adjusting with the 10 solution,
> and the 4,10 knob, as you move the reading from 9.something, towards the
> 10, the 4 will be moving to 4.something. So if I have to choose between
> the accuracy on the 10 side or the accuracy on the 4 side, then I will
> chose based on whether I am using that meter for low Ph fresh water, or
> for salt water. A new probe is not likely to be very inaccurate, but as
> they age, they start loosing some accuracy.
>
> Now if you were to purchase a Ph meter that only had one calibration
> knob on it, which calibration knob would you like to be without. I would
> not want to be without either one :-)
>
> Wayne Sallee
> Wayne's Pets
>
>
> Wayne Sallee wrote on 5/2/2007 4:23 PM:
>> I don't care for the design of the Miwukee. It's more toyish in
>> design, and does not have the same calibration options as a better
>> quality ph meter.
>>
>> Wayne Sallee
>> Wayne's Pets
>>
>>
>> Pszemol wrote on 5/2/2007 7:54 AM:
>>> "Wayne Sallee" > wrote in message
>>> nk.net...
>>>> bo00_ger wrote on 4/28/2007 1:43 PM:
>>>>> thinking about buying a pH meter for continuous operation in a sump.
>>>>> Which one to get? Pinpoint? Milwaukee?
>>>>
>>>> Pinpoint.
>>>
>>> Would you care to give any particular *reason* for this choice?
Pszemol
May 3rd 07, 06:18 AM
"Wayne Sallee" > wrote in message link.net...
> A Little More Info About Ph Meters:
>
> Most people calibrate ph meters with two calibration
> fluids. I use all 3 (4,7,10). Yes you can calibrate
> with just two, but I'm a perfectionist :-)
Wayne - software in your meter does not use 3 points.
You are wasting your calibration fluids doing pH 4
calibration for marine measurements and pH 10 fluid
in freshwater measurements.
Software is designed for acurate measuring
BETWEEN calibration points. So - when your
sample is between 7 an 10, like for marine
environments - calibration point at pH 4 is not
adding you accuracy.
You are doing more harm than help adjusting
your pH meter for acurate pH 4 when you measure
marine water.
> Now here's why:
>
> Imagine you have a yard stick, or a meter stick. Now
> imagine that 4 is on the far left, and 10 is on the
> far right, and 7 is in the middle. Now with the 7
> knob what you are doing is moving that measuring
> stick left or right to make 7 line up in the middle
> where the 7 is supposed to be. When you calibrate,
> you should do the 7 first.
>
> Now imagine that you are stretching and compressing
> the measuring stick so that the 7 and 10 line up.
> That's what the other knob does.
This is the problem with your thinking:
pH probe is not ideal and is not linear! The calibration
curve will need a DIFFERENT correction above pH 7
and below pH 7. THAT IS THE REASON YOU DO
CALIBRATION WRAPING YOUR EXPECTED SAMPLE pH.
> So let's say you
> are calibrating with the 7,10 knob, with a 10
> solution, and let's say it reads 9.something. And
> then you check the 4 with a 4 solution, and it reads
> 4.0. As you are adjusting with the 10 solution, and
> the 4,10 knob, as you move the reading from
> 9.something, towards the 10, the 4 will be moving to
> 3.something. So if I have to choose between the
> accuracy on the 10 side or the accuracy on the 4
> side, then I will chose based on whether I am using
> that meter for low Ph fresh water, or for salt
> water. A new probe is not likely to be very
> inaccurate, but as they age, they start loosing some
> accuracy.
The problem is that when you adjust reading at pH 4
you influence reading at pH 10. Each range will have
conflicting needs and calibration point should always
wrap around the expected sample pH.
> Now if you were to purchase a Ph meter that only had
> one calibration knob on it, which calibration knob
> would you like to be without. I would not want to be
> without either one :-)
Calibration at pH 7 is in fact compensating "zero point"
for the pH probe and the whole measuring equipment.
It is more important than pH 4 or 10 calibration points,
which is setting only the DC amplifier gain - this varies
less from probe to probe and can be preset at factory
in cheaper, less precise units.
Mr.pH
May 3rd 07, 11:13 AM
Just in case:
http://www.ph-meter.info/pH-electrode-calibration
Borek AKA Mr.pH
--
http://www.chembuddy.com/?left=BATE&right=pH-calculator
http://www.ph-meter.info/pH-electrode
http://www.bpp.com.pl/?left=dysleksja&right=dysleksja
http://www.terapia-kregoslupa.waw.pl
Wayne Sallee
May 3rd 07, 06:17 PM
Mr.pH wrote on 5/3/2007 5:13 AM:
> http://www.bpp.com.pl/?left=dysleksja&right=dysleksja
> http://www.terapia-kregoslupa.waw.pl
>
Well what a coincidink :-) It's in Polish :-)
Now who here speaks Polish besides Pszemol?
Now Pszemol if you wanted to post some links to web
sites you could have used your own handle, instead
of making it look like you were someone else.
Wayne Sallee
Wayne's Pets
Wayne Sallee
May 3rd 07, 06:27 PM
Pszemol wrote on 5/3/2007 12:18 AM:
> "Wayne Sallee" > wrote in message
> link.net...
>> A Little More Info About Ph Meters:
>>
>> Most people calibrate ph meters with two calibration fluids. I use all
>> 3 (4,7,10). Yes you can calibrate with just two, but I'm a
>> perfectionist :-)
>
> Wayne - software in your meter does not use 3 points.
> You are wasting your calibration fluids doing pH 4
> calibration for marine measurements and pH 10 fluid
> in freshwater measurements.
I did not write the software for the Ph meters, but
I do know what the results are when I use my
calibration methods, so I do know what the end
results are. And what is most important here, is
what the end results are.
>
> Software is designed for acurate measuring
> BETWEEN calibration points. So - when your
> sample is between 7 an 10, like for marine
> environments - calibration point at pH 4 is not
> adding you accuracy.
Most people calibrate their Ph meters for us in salt
water with 7, and 4. I was stating that it is better
to calibrate with 10. I also stated that I like to
look at 4 as well, and see where it is. And if I use
a ph meter (I have more than one ph meter) mainly
for fresh water I will go with the 4 ph calibration
as the standard, but still like to look at the 10.
> You are doing more harm than help adjusting
> your pH meter for acurate pH 4 when you measure
> marine water.
That's silly. I am not doing more harm than good by
using all 3 calibration fluids.
>
>> Now here's why:
>>
>> Imagine you have a yard stick, or a meter stick. Now imagine that 4 is
>> on the far left, and 10 is on the far right, and 7 is in the middle.
>> Now with the 7 knob what you are doing is moving that measuring stick
>> left or right to make 7 line up in the middle where the 7 is supposed
>> to be. When you calibrate, you should do the 7 first.
>>
>> Now imagine that you are stretching and compressing the measuring
>> stick so that the 7 and 10 line up. That's what the other knob does.
>
> This is the problem with your thinking:
> pH probe is not ideal and is not linear! The calibration
> curve will need a DIFFERENT correction above pH 7
> and below pH 7. THAT IS THE REASON YOU DO
> CALIBRATION WRAPING YOUR EXPECTED SAMPLE pH.
I don't understand what you are trying to say here.
>
>> So let's say you are calibrating with the 7,10 knob, with a 10
>> solution, and let's say it reads 9.something. And then you check the 4
>> with a 4 solution, and it reads 4.0. As you are adjusting with the 10
>> solution, and the 4,10 knob, as you move the reading from 9.something,
>> towards the 10, the 4 will be moving to 3.something. So if I have to
>> choose between the accuracy on the 10 side or the accuracy on the 4
>> side, then I will chose based on whether I am using that meter for low
>> Ph fresh water, or for salt water. A new probe is not likely to be
>> very inaccurate, but as they age, they start loosing some accuracy.
>
> The problem is that when you adjust reading at pH 4
> you influence reading at pH 10. Each range will have
> conflicting needs and calibration point should always
> wrap around the expected sample pH.
You need to re-read what I said.
>
>> Now if you were to purchase a Ph meter that only had one calibration
>> knob on it, which calibration knob would you like to be without. I
>> would not want to be without either one :-)
>
> Calibration at pH 7 is in fact compensating "zero point"
> for the pH probe and the whole measuring equipment.
> It is more important than pH 4 or 10 calibration points,
> which is setting only the DC amplifier gain - this varies
> less from probe to probe and can be preset at factory
> in cheaper, less precise units.
Yes, 7 is the most important point. That's why it
should be set first. So wha'ts your point?
Wayne Sallee
Wayne's Pets
Wayne Sallee
May 3rd 07, 06:56 PM
Wayne Sallee wrote on 5/3/2007 12:27 PM:
>
> Most people calibrate their Ph meters for us in salt water with 7, and
> 4.
oops, That was supposed to read "meters for use in
salt water".
Wayne Sallee
Wayne's Pets
Borek
May 3rd 07, 10:58 PM
On Thu, 03 May 2007 19:17:17 +0200, Wayne Sallee >
wrote:
>> http://www.bpp.com.pl/?left=dysleksja&right=dysleksja
>> http://www.terapia-kregoslupa.waw.pl
>
> Well what a coincidink :-) It's in Polish :-)
>
> Now who here speaks Polish besides Pszemol?
>
> Now Pszemol if you wanted to post some links to web sites you could have
> used your own handle, instead of making it look like you were someone
> else.
While Pszemol's nick suggests he is Polish (at least for someone that
speaks Polish) he is not the only Polish posting on usenet. And I am
definitely not him ;)
Borek AKA Mr.pH
--
http://www.chembuddy.com/?left=BATE&right=pH-calculator
http://www.ph-meter.info/pH-electrode
http://www.bpp.com.pl/?left=dysleksja&right=dysleksja
http://www.terapia-kregoslupa.waw.pl
Pszemol
May 5th 07, 01:49 PM
"Wayne Sallee" > wrote in message ink.net...
> Mr.pH wrote on 5/3/2007 5:13 AM:
>> http://www.bpp.com.pl/?left=dysleksja&right=dysleksja
>> http://www.terapia-kregoslupa.waw.pl
>>
>
> Well what a coincidink :-) It's in Polish :-)
>
> Now who here speaks Polish besides Pszemol?
>
> Now Pszemol if you wanted to post some links to web
> sites you could have used your own handle, instead
> of making it look like you were someone else.
Wayne - stop being paranoic... "Mr.pH" is not me :-)
There is about 38 milions residents in Poland alone
and about 10 milions somewhere around the World.
Including 2-3 milions (or more?) immigrants in USA.
The chances are - more than one of them is here in our group.
And your "coincidink" is not a good proof of anything...
I hope your are not often chosen to do jury duty, Wayne! ;-)
Pszemol
May 5th 07, 03:20 PM
"Wayne Sallee" > wrote in message nk.net...
> I did not write the software for the Ph meters, but
> I do know what the results are when I use my
> calibration methods, so I do know what the end
> results are. And what is most important here, is
> what the end results are.
Repeat after me:
calibration at pH 4 is not needed for sal****er measurements.
Only pH 9 or 10 makes sense in relation to zeroing in pH 7.0
> Most people calibrate their Ph meters for us in salt
> water with 7, and 4.
Only if they do not read the instructions included with pHmeter.
> I was stating that it is better to calibrate with 10.
You were correct in this statement.
But you were wrong suggesting that additional calibration at pH 4
improves measurements between pH 7 and 10. It does not.
If anything - it might decrease pH meter accuracy in the 7-10 range.
> I also stated that I like to look at 4 as well, and see where it is.
> And if I use a ph meter (I have more than one ph meter) mainly
> for fresh water I will go with the 4 ph calibration
> as the standard, but still like to look at the 10.
Let me asure you that pH meter uses these ranges SEPARATELLY.
>> You are doing more harm than help adjusting
>> your pH meter for acurate pH 4 when you measure
>> marine water.
>
> That's silly. I am not doing more harm than good by
> using all 3 calibration fluids.
Yes, you are.
Calibration at pH 4 influences corrections you made for pH 10.
These are two separate ranges your pH meter operates and
it usually does not have a memory to keep separate settings
for both ranges even if the calibration is automatic/digital.
It is most certainly bad, if the meter has a turn knob (trimmer
setting the amplifier gain directly with the variable resistance).
>> This is the problem with your thinking:
>> pH probe is not ideal and is not linear! The calibration
>> curve will need a DIFFERENT correction above pH 7
>> and below pH 7. THAT IS THE REASON YOU DO
>> CALIBRATION WRAPING YOUR EXPECTED SAMPLE pH.
>
> I don't understand what you are trying to say here.
Ask the question... I am electronic engineer involved in
design of electronic circuits - I will try to explain as good as I can... :-)
>> The problem is that when you adjust reading at pH 4
>> you influence reading at pH 10. Each range will have
>> conflicting needs and calibration point should always
>> wrap around the expected sample pH.
>
> You need to re-read what I said.
When you calibrate at pH 4 you DISTURB calibration made for pH 10.
>>> Now if you were to purchase a Ph meter that only had one calibration
>>> knob on it, which calibration knob would you like to be without. I
>>> would not want to be without either one :-)
>>
>> Calibration at pH 7 is in fact compensating "zero point"
>> for the pH probe and the whole measuring equipment.
>> It is more important than pH 4 or 10 calibration points,
>> which is setting only the DC amplifier gain - this varies
>> less from probe to probe and can be preset at factory
>> in cheaper, less precise units.
>
>
> Yes, 7 is the most important point. That's why it
> should be set first. So wha'ts your point?
My point is that only very expensive, lab-grade pH meters costing
well above $1000 have ability to keep multipoint calibration curves
for measurements in wide ranges of sample pH. Our regular meters
can keep only one curve in memory (if digital) or have fixed aplifier
gain set with the "knob" (which is a variable resistor in a feedback
loop of the operational amplifier). Due to the nature of the beast,
you should always calibrate with the buffors reflecting your expected
sample pH the way, sample pH lays inbetween the calibration points.
Calibration for pH 4 made after calibration at pH 10 does not make
sense if you measure pH 8.3 and it affects previously made
calibration for pH 10. In effect makes your meter LESS accurate
for measurements in the pH 7-10 range.
Wayne Sallee
May 5th 07, 06:25 PM
Pszemol wrote on 5/5/2007 9:20 AM:
> "Wayne Sallee" > wrote in message
> nk.net...
>> I did not write the software for the Ph meters, but I do know what the
>> results are when I use my calibration methods, so I do know what the
>> end results are. And what is most important here, is what the end
>> results are.
>
> Repeat after me:
> calibration at pH 4 is not needed for sal****er measurements. Only pH 9
> or 10 makes sense in relation to zeroing in pH 7.0
>
>> Most people calibrate their Ph meters for us in salt water with 7, and 4.
>
> Only if they do not read the instructions included with pHmeter.
>
>> I was stating that it is better to calibrate with 10.
>
> You were correct in this statement. But you were wrong suggesting that
> additional calibration at pH 4
> improves measurements between pH 7 and 10. It does not.
> If anything - it might decrease pH meter accuracy in the 7-10 range.
>
It helps show degradation of the accuracy of the
probe as it gets older.
>> I also stated that I like to look at 4 as well, and see where it is.
>> And if I use a ph meter (I have more than one ph meter) mainly for
>> fresh water I will go with the 4 ph calibration as the standard, but
>> still like to look at the 10.
>
> Let me asure you that pH meter uses these ranges SEPARATELLY.
>
No they don't. Not the ones that I've used.
The Ph 4,10 knob adjusts both 4 and 10 ends of the
scale a the same time.
>>> You are doing more harm than help adjusting
>>> your pH meter for acurate pH 4 when you measure
>>> marine water.
>>
>> That's silly. I am not doing more harm than good by using all 3
>> calibration fluids.
>
> Yes, you are.
> Calibration at pH 4 influences corrections you made for pH 10.
> These are two separate ranges your pH meter operates and
> it usually does not have a memory to keep separate settings
> for both ranges even if the calibration is automatic/digital. It is most
> certainly bad, if the meter has a turn knob (trimmer
> setting the amplifier gain directly with the variable resistance).
>
Like I said in a few lines up, the Ph 4,10 knob
adjusts both 4 and 10 ends of the scale a the same time.
>>> This is the problem with your thinking:
>>> pH probe is not ideal and is not linear! The calibration
>>> curve will need a DIFFERENT correction above pH 7
>>> and below pH 7. THAT IS THE REASON YOU DO
>>> CALIBRATION WRAPING YOUR EXPECTED SAMPLE pH.
>>
>> I don't understand what you are trying to say here.
>
> Ask the question... I am electronic engineer involved in
> design of electronic circuits - I will try to explain as good as I
> can... :-)
And I have been involved with writting machine code
to operate computer circuits, but what does this
have to do with the circuits in the ph probes? What
experience have you had with Ph probe circuits?
>
>>> The problem is that when you adjust reading at pH 4
>>> you influence reading at pH 10. Each range will have
>>> conflicting needs and calibration point should always
>>> wrap around the expected sample pH.
>>
>> You need to re-read what I said.
>
> When you calibrate at pH 4 you DISTURB calibration made for pH 10.
>
Exactly my point. You are adjusting both the 4 and
the 10 end of the scale when you adjust one of them.
Of course you are also affecting the whole scale.
You can look a this as being like compressing or
stretching a meter stick.
>>>> Now if you were to purchase a Ph meter that only had one calibration
>>>> knob on it, which calibration knob would you like to be without. I
>>>> would not want to be without either one :-)
>>>
>>> Calibration at pH 7 is in fact compensating "zero point"
>>> for the pH probe and the whole measuring equipment.
>>> It is more important than pH 4 or 10 calibration points,
>>> which is setting only the DC amplifier gain - this varies
>>> less from probe to probe and can be preset at factory
>>> in cheaper, less precise units.
>>
>>
>> Yes, 7 is the most important point. That's why it should be set first.
>> So wha'ts your point?
>
> My point is that only very expensive, lab-grade pH meters costing
> well above $1000 have ability to keep multipoint calibration curves
> for measurements in wide ranges of sample pH. Our regular meters
> can keep only one curve in memory (if digital) or have fixed aplifier
> gain set with the "knob" (which is a variable resistor in a feedback
> loop of the operational amplifier). Due to the nature of the beast,
> you should always calibrate with the buffors reflecting your expected
> sample pH the way, sample pH lays inbetween the calibration points.
> Calibration for pH 4 made after calibration at pH 10 does not make
> sense if you measure pH 8.3 and it affects previously made
> calibration for pH 10. In effect makes your meter LESS accurate
> for measurements in the pH 7-10 range.
It does make sence if you are going to be using your
ph meter to measure fluids from 5.5 to 8.5
When I calibrate my ph meters, I like to know how
they are going to read across the entire scale.
Wayne Sallee
Wayne's Pets
Pszemol
May 6th 07, 03:04 PM
"Wayne Sallee" > wrote in message ink.net...
>>> I was stating that it is better to calibrate with 10.
>>
>> You were correct in this statement. But you were wrong suggesting that additional calibration at pH 4
>> improves measurements between pH 7 and 10. It does not.
>> If anything - it might decrease pH meter accuracy in the 7-10 range.
>>
>
> It helps show degradation of the accuracy of the probe as it gets older.
not really... It only shows how different characteristic of the probe
is in these two ranges: 7-10 and 4-7.
>>> I also stated that I like to look at 4 as well, and see where it is.
>>> And if I use a ph meter (I have more than one ph meter) mainly for fresh water I will go with the 4 ph calibration as the
>>> standard, but still like to look at the 10.
>>
>> Let me asure you that pH meter uses these ranges SEPARATELLY.
>>
>
> No they don't. Not the ones that I've used.
> The Ph 4,10 knob adjusts both 4 and 10 ends of the scale a the same time.
That is exactly what I said: this meter uses one setting for two separate ranges. If you pick one setting of this knob good for pH 4
it will not be good for pH 10 and vice versa. If you get something between, averaging two calibrations, you will get both ranges not
accurate as good as they could be.
From the meter perspective both ranges are separated and it does not have intelligence to connect them anyhow. They are separated by
the probe characteristic which is different for each range and is aproximated with linear at different angle for each measurement
range.
> Like I said in a few lines up, the Ph 4,10 knob adjusts both 4 and 10 ends of the scale a the same time.
pH 4/10 knob is ONE, and is responsible for ONE setting of operational amplifier gain.
This sets the angle for the linear aproximation of non-linear probe characteristic in one of these two ranges.
>>> I don't understand what you are trying to say here.
>>
>> Ask the question... I am electronic engineer involved in
>> design of electronic circuits - I will try to explain as good as I can... :-)
>
> And I have been involved with writting machine code to operate computer circuits, but what does this have to do with the circuits
> in the ph probes?
It has a lot to do with circuits in pH meters.
The measuring principle is the same in many other elecronic meters.
You need to set the zero point and then adjust angle of the linear
aproximation - this second setting is just gain of the amplifier...
output voltage = gain(input voltage) + offset.
or
output voltage = gain(input voltage + offset)
depending on the circuit.
pH probe is just weak source of electric DC voltage with very high internal resistance ranging 100-200 megaohm.
Your pH meter is just very precise DC volt meter with adjusted gain the way that miliVolts are translated to pH readings.
> What experience have you had with Ph probe circuits?
Well, I have build one simple pH meter myself using one LF444 integrated circuit containing 4 opamps.
If you have basic electronic skills and know how to read and understand electronic schematics you can try to understand how the pH
meter works from the LF444 datasheet. Manufacturer of this simple chip is showing its usage in measuring devices using pH meter guts
as an example:
http://cache.national.com/ds/LF/LF444.pdf
Check out the page 9 - you will find two adjustable resistors, "knobs" like you call them, one is marked "calibration" and this one
is adjusting zero point of the amplifier (pH 7 adjust) and the other, marked as "temperature" adjusts gain of the amplifier, for the
user this will be your "pH 4/10" adjust.
Output from the circuit is just variable voltage which can be read with any digital or analog voltmeter. Your pH meter will have
both parts in one box: voltmeter with digital readout and the preamplifier with adjustments to prepare DC signal generated by the pH
probe to be read by the voltmeter.
BTW - there is no electronics in the probe itself. Probe is a pure electrochemical devices with chemical processes causing voltage
difference measured by the high imput resistance operational amplifiers. Sometimes probe is embedded with a temperature sensor in a
form of a thermistor or PT100 thermopair sensor.
As you can see, adjustment is one for pH 4 as for pH 10, so if the probe characteristic in these two ranges differs significantly,
you better adjust the meter for only one range at the time, take the measurements and than if second sample is in a different range,
calibrate again in the range reflecting the measured sample.
>>>> The problem is that when you adjust reading at pH 4
>>>> you influence reading at pH 10. Each range will have
>>>> conflicting needs and calibration point should always
>>>> wrap around the expected sample pH.
>>>
>>> You need to re-read what I said.
>>
>> When you calibrate at pH 4 you DISTURB calibration made for pH 10.
>>
>
> Exactly my point. You are adjusting both the 4 and the 10 end of the scale when you adjust one of them. Of course you are also
> affecting the whole scale. You can look a this as being like compressing or stretching a meter stick.
Dont say exactly my point because you were suggesting calibration of the meter with two solutions,
which is wrong. You will have better calibration results for given range if you calibrate with only calibration solution.
Calibration with pH 4 disturbs previous calibration made with pH 10 and meter will NOT BE AS ACURATE for measurements in sal****er
as it would be if calibrated with only pH 10.
> It does make sence if you are going to be using your ph meter to measure fluids from 5.5 to 8.5
>
> When I calibrate my ph meters, I like to know how they are going to read across the entire scale.
The best option is to calibrate meter with pH 4 for measurements of fluid of pH 4-7 and to calibrate with pH 10 formeasurements of
fluids of pH 7-10.
As you will see in above discussion calibration with pH 4 makes readings in 7-10 range LESS ACURATE. How much less? It depends on
specific probe you use - each probe will be different.
Wayne Sallee
May 6th 07, 08:42 PM
Pszemol wrote on 5/6/2007 9:04 AM:
> "Wayne Sallee" > wrote in message
> ink.net...
>>>> I was stating that it is better to calibrate with 10.
>>>
>>> You were correct in this statement. But you were wrong suggesting
>>> that additional calibration at pH 4
>>> improves measurements between pH 7 and 10. It does not.
>>> If anything - it might decrease pH meter accuracy in the 7-10 range.
>>>
>>
>> It helps show degradation of the accuracy of the probe as it gets older.
>
> not really... It only shows how different characteristic of the probe
> is in these two ranges: 7-10 and 4-7.
>
>>>> I also stated that I like to look at 4 as well, and see where it is.
>>>> And if I use a ph meter (I have more than one ph meter) mainly for
>>>> fresh water I will go with the 4 ph calibration as the standard, but
>>>> still like to look at the 10.
>>>
>>> Let me asure you that pH meter uses these ranges SEPARATELLY.
>>>
>>
>> No they don't. Not the ones that I've used.
>> The Ph 4,10 knob adjusts both 4 and 10 ends of the scale a the same time.
>
> That is exactly what I said: this meter uses one setting for two
> separate ranges. If you pick one setting of this knob good for pH 4 it
> will not be good for pH 10 and vice versa. If you get something between,
> averaging two calibrations, you will get both ranges not accurate as
> good as they could be.
> From the meter perspective both ranges are separated and it does not
> have intelligence to connect them anyhow. They are separated by the
> probe characteristic which is different for each range and is
> aproximated with linear at different angle for each measurement range.
>
Well of course, if they were not that way, then it
would not be as accurate. And some Ph meters have
been made with a knob for 7, a knob for 4, and
another knob for 10. So what's your point?
>> Like I said in a few lines up, the Ph 4,10 knob adjusts both 4 and 10
>> ends of the scale a the same time.
>
> pH 4/10 knob is ONE, and is responsible for ONE setting of operational
> amplifier gain.
> This sets the angle for the linear aproximation of non-linear probe
> characteristic in one of these two ranges.
>
No. You can check the calibration of a ph meter, and
have 10 and 4 dead on, and then recalibrate the 10
to 10.something, then go to the 4, with a 4
solution, and then bring it back to 4, and the 10
will be back to 10.
>>>> I don't understand what you are trying to say here.
>>>
>>> Ask the question... I am electronic engineer involved in
>>> design of electronic circuits - I will try to explain as good as I
>>> can... :-)
>>
>> And I have been involved with writting machine code to operate
>> computer circuits, but what does this have to do with the circuits in
>> the ph probes?
>
> It has a lot to do with circuits in pH meters.
> The measuring principle is the same in many other elecronic meters.
> You need to set the zero point and then adjust angle of the linear
> aproximation - this second setting is just gain of the amplifier...
>
> output voltage = gain(input voltage) + offset.
> or
> output voltage = gain(input voltage + offset)
> depending on the circuit.
>
> pH probe is just weak source of electric DC voltage with very high
> internal resistance ranging 100-200 megaohm.
> Your pH meter is just very precise DC volt meter with adjusted gain the
> way that miliVolts are translated to pH readings.
>
>> What experience have you had with Ph probe circuits?
>
> Well, I have build one simple pH meter myself using one LF444 integrated
> circuit containing 4 opamps.
> If you have basic electronic skills and know how to read and understand
> electronic schematics you can try to understand how the pH meter works
> from the LF444 datasheet. Manufacturer of this simple chip is showing
> its usage in measuring devices using pH meter guts as an example:
> http://cache.national.com/ds/LF/LF444.pdf
>
> Check out the page 9 - you will find two adjustable resistors, "knobs"
> like you call them,
Yea and I could have also called them trimmers, but
most people would not have known what I was talking
about if I did that.
> one is marked "calibration" and this one is
> adjusting zero point of the amplifier (pH 7 adjust) and the other,
> marked as "temperature" adjusts gain of the amplifier, for the user this
> will be your "pH 4/10" adjust.
It looks to me like this design was not designed for
a 4,10 adjustment.
The instructions says:
"To calibrate, insert probe in pH =7 solution. Set
the “TEMPERATURE ADJUST” pot, R2, to correspond to
the solution temperature: full clockwise for 0°C, and
proportionately for intermediate temperatures, using
a turns-counting dial. Then set “CALIBRATE” pot so
output reads 7V."
By the way, the pinpoint probes do automatically
adjust for temp.
> Output from the circuit is just variable voltage which can be read with
> any digital or analog voltmeter. Your pH meter will have both parts in
> one box: voltmeter with digital readout and the preamplifier with
> adjustments to prepare DC signal generated by the pH probe to be read by
> the voltmeter.
> BTW - there is no electronics in the probe itself. Probe is a pure
> electrochemical devices with chemical processes causing voltage
> difference measured by the high imput resistance operational amplifiers.
> Sometimes probe is embedded with a temperature sensor in a form of a
> thermistor or PT100 thermopair sensor.
>
> As you can see, adjustment is one for pH 4 as for pH 10, so if the probe
> characteristic in these two ranges differs significantly, you better
> adjust the meter for only one range at the time, take the measurements
> and than if second sample is in a different range, calibrate again in
> the range reflecting the measured sample.
>
>>>>> The problem is that when you adjust reading at pH 4
>>>>> you influence reading at pH 10. Each range will have
>>>>> conflicting needs and calibration point should always
>>>>> wrap around the expected sample pH.
>>>>
>>>> You need to re-read what I said.
>>>
>>> When you calibrate at pH 4 you DISTURB calibration made for pH 10.
>>>
>>
>> Exactly my point. You are adjusting both the 4 and the 10 end of the
>> scale when you adjust one of them. Of course you are also affecting
>> the whole scale. You can look a this as being like compressing or
>> stretching a meter stick.
>
> Dont say exactly my point because you were suggesting calibration of the
> meter with two solutions,
> which is wrong.
No I was saying that I like to calibrate with 3
solutions.
> You will have better calibration results for given range
> if you calibrate with only calibration solution.
> Calibration with pH 4 disturbs previous calibration made with pH 10 and
> meter will NOT BE AS ACURATE for measurements in sal****er as it would
> be if calibrated with only pH 10.
>
>> It does make sence if you are going to be using your ph meter to
>> measure fluids from 5.5 to 8.5
>>
>> When I calibrate my ph meters, I like to know how they are going to
>> read across the entire scale.
>
> The best option is to calibrate meter with pH 4 for measurements of
> fluid of pH 4-7 and to calibrate with pH 10 formeasurements of fluids of
> pH 7-10.
> As you will see in above discussion calibration with pH 4 makes readings
> in 7-10 range LESS ACURATE. How much less? It depends on specific probe
> you use - each probe will be different.
Not everybody uses their Ph probe for only 7 to 10,
or 4 to 7.
Wayne Sallee
Wayne's Pets
Borek
May 7th 07, 08:26 AM
On Sun, 06 May 2007 16:04:24 +0200, Pszemol > wrote:
> From the meter perspective both ranges are separated and it does not
> have intelligence to connect them anyhow. They are separated by the
> probe characteristic which is different for each range and is
> aproximated with linear at different angle for each measurement range.
Note that theoretical response IS linear, so calibration means only
adjusting for the lack of perfectness ;)
> Well, I have build one simple pH meter myself using one LF444 integrated
> circuit containing 4 opamps.
> If you have basic electronic skills and know how to read and understand
> electronic schematics you can try to understand how the pH meter works
> from the LF444 datasheet. Manufacturer of this simple chip is showing
> its usage in measuring devices using pH meter guts as an example:
> http://cache.national.com/ds/LF/LF444.pdf
See other example at
http://www.ph-meter.info/pH-meter-construction
Two calibration resistors called - not surprisingly - slope and zero :)
Mr.pH
--
http://www.chembuddy.com/?left=BATE&right=pH-calculator
http://www.ph-meter.info/pH-electrode
http://www.bpp.com.pl/?left=dysleksja&right=dysleksja
http://www.terapia-kregoslupa.waw.pl
Borek
May 7th 07, 08:48 AM
On Sun, 06 May 2007 21:42:58 +0200, Wayne Sallee >
wrote:
> No. You can check the calibration of a ph meter, and have 10 and 4 dead
> on, and then recalibrate the 10 to 10.something, then go to the 4, with
> a 4 solution, and then bring it back to 4, and the 10 will be back to 10.
You said it by yourself: you calibrate at 10.00, then you move electrode
to 4.00 buffer, it shows 4.something, so you calibrate it to show 4.00,
now you go back to 10.00 buffer - and your indications are no longer 10.00
but 10.-something. That's exactly Pszemol's point and you have just
confirmed he is right :)
I will stick with two-point calibration, depending on the range I am
interested in. Above 7 - calibrate for 7/10. Below 7 - calibrate for 4/7..
You want to register titration curve and you need 2-11 range - I am in
troubles ;)
Mr.pH
--
http://www.chembuddy.com/?left=BATE&right=pH-calculator
http://www.ph-meter.info/pH-electrode
http://www.bpp.com.pl/?left=dysleksja&right=dysleksja
http://www.terapia-kregoslupa.waw.pl
Wayne Sallee
May 7th 07, 03:01 PM
Borek wrote on 5/7/2007 2:48 AM:
> On Sun, 06 May 2007 21:42:58 +0200, Wayne Sallee >
> wrote:
>
>> No. You can check the calibration of a ph meter, and have 10 and 4
>> dead on, and then recalibrate the 10 to 10.something, then go to the
>> 4, with a 4 solution, and then bring it back to 4, and the 10 will be
>> back to 10.
>
> You said it by yourself: you calibrate at 10.00, then you move electrode
> to 4.00 buffer, it shows 4.something, so you calibrate it to show 4.00,
> now you go back to 10.00 buffer - and your indications are no longer
> 10.00 but 10.-something. That's exactly Pszemol's point and you have
> just confirmed he is right :)
Don't twist my words. That's not what I said.
Wayne Sallee
Wayne's Pets
Borek
May 8th 07, 01:11 AM
On Mon, 07 May 2007 16:01:50 +0200, Wayne Sallee >
wrote:
>>> No. You can check the calibration of a ph meter, and have 10 and 4
>>> dead on, and then recalibrate the 10 to 10.something, then go to the
>>> 4, with a 4 solution, and then bring it back to 4, and the 10 will be
>>> back to 10.
>> You said it by yourself: you calibrate at 10.00, then you move
>> electrode to 4.00 buffer, it shows 4.something, so you calibrate it to
>> show 4.00, now you go back to 10.00 buffer - and your indications are
>> no longer 10.00 but 10.-something. That's exactly Pszemol's point and
>> you have just confirmed he is right :)
>
> Don't twist my words. That's not what I said.
These are not your exact words, but you have described how calibration at
10 can be changed by calibration at 4. What I wrote is just an obvious and
logical conclusion. When you calibrate at 4 you change calibration at 10
(your words), thus if planning to measure pH in the 7-10 range you better
stick to two point (7 and 10) calibration (conclusion).
Mr.pH
--
http://www.chembuddy.com/?left=BATE&right=pH-calculator
http://www.ph-meter.info/pH-electrode
http://www.bpp.com.pl/?left=dysleksja&right=dysleksja
http://www.terapia-kregoslupa.waw.pl
Pszemol
May 8th 07, 04:20 PM
"Borek" > wrote in message news:op.trzetizvu0t02r@borek...
> These are not your exact words, but you have described how calibration at
> 10 can be changed by calibration at 4. What I wrote is just an obvious and
> logical conclusion. When you calibrate at 4 you change calibration at 10
> (your words), thus if planning to measure pH in the 7-10 range you better
> stick to two point (7 and 10) calibration (conclusion).
Fully agree.
Wayne, have you inspected the electronic schematic using LF444 chip?
After you take a look at the schematic everything should be clearer ;-)
Pszemol
May 8th 07, 04:25 PM
"Borek" > wrote in message news:op.trx4avsiu0t02r@borek...
> > From the meter perspective both ranges are separated and it does not
> > have intelligence to connect them anyhow. They are separated by the
> > probe characteristic which is different for each range and is
> > aproximated with linear at different angle for each measurement range.
>
> Note that theoretical response IS linear, so calibration means only
> adjusting for the lack of perfectness ;)
Yes, and if you have new and good quality pH probe you will probably
be able to set the pH meter to calibrate good enough in both points
(4 and 10). But when the time goes on and probe characteristic drifts
from the ideal linerality both parts of the measuring range become
more and more appart from each other and probe requires then separate
calibration for samples from opposite ranges.
> See other example at
> http://www.ph-meter.info/pH-meter-construction
>
> Two calibration resistors called - not surprisingly - slope and zero :)
Much simpler schematic, easier to understand even for a layman.
Wayne Sallee
May 10th 07, 02:17 PM
Yes I do know how to read a schematic.
Wayne Sallee
Wayne's Pets
Pszemol wrote on 5/8/2007 10:20 AM:
> "Borek" > wrote in
> message news:op.trzetizvu0t02r@borek...
>> These are not your exact words, but you have described how calibration
>> at 10 can be changed by calibration at 4. What I wrote is just an
>> obvious and logical conclusion. When you calibrate at 4 you change
>> calibration at 10 (your words), thus if planning to measure pH in the
>> 7-10 range you better stick to two point (7 and 10) calibration
>> (conclusion).
>
> Fully agree.
>
> Wayne, have you inspected the electronic schematic using LF444 chip?
> After you take a look at the schematic everything should be clearer ;-)
daneyalhardy
February 14th 11, 04:49 PM
At pH 7 calibration compensation is actually "zero" in the pH probe and the measurement equipment. More important than the correction of pH 4 or 10 points this is the only set the DC gain of the amplifier - this change less from the sensor probe can be preset at the factory in the cheap, not accurate enough units.