It's a month since I started using a permanent pH probe to control my
tank's CO2 injections. The probe being used is made by AquaMedic. At the
the first calibration I used buffer solutions 7 and 10. The instructions
say to use 7 and 9. I used the 10 instead to adjust the calibration
curve. The calibrated probe's readings matched the colorimeter's readings.

It's a month later. Time to recalibrate. I first checked to see how far
the probe had drifted. It was less than .1 according to the buffer
solutions. In this calibration routine I used 7 and 9.18. Calibration
done I then placed the probe into the tank water. It read 6.85. The last
reading before calibration was 7.25. I now checked with the colorimeter
which said the tank was 7.3. The probe is now off by -0.4!

I recalibrated with 7 and the original 10. Still reads 6.85!

It seems to me that I have somehow trashed the probe by recalibrating
it. Is there anything I can do to recover the probe? A temporary
solution I have, before I can get a replacement, is to calibrate the
probe to the known tank pH. That is, placing the probe in the tank water
and adjusting it until it reads around about 7.3.

Anyone have a similar experience?

Nikki

  #2  
Old January 21st 05, 10:05 PM
Rick
external usenet poster
 
Posts: n/a
Default


"Nikki Casali" wrote in message
...
It's a month since I started using a permanent pH probe to control my
tank's CO2 injections. The probe being used is made by AquaMedic. At the
the first calibration I used buffer solutions 7 and 10. The instructions
say to use 7 and 9. I used the 10 instead to adjust the calibration
curve. The calibrated probe's readings matched the colorimeter's readings.

It's a month later. Time to recalibrate. I first checked to see how far
the probe had drifted. It was less than .1 according to the buffer
solutions. In this calibration routine I used 7 and 9.18. Calibration
done I then placed the probe into the tank water. It read 6.85. The last
reading before calibration was 7.25. I now checked with the colorimeter
which said the tank was 7.3. The probe is now off by -0.4!

I recalibrated with 7 and the original 10. Still reads 6.85!

It seems to me that I have somehow trashed the probe by recalibrating
it. Is there anything I can do to recover the probe? A temporary
solution I have, before I can get a replacement, is to calibrate the
probe to the known tank pH. That is, placing the probe in the tank water
and adjusting it until it reads around about 7.3.

Anyone have a similar experience?

Nikki


without using the calibration mode simply check the PH level of the buffer
you are using. You are then using a known solution. If the meter reads
correctly then your water PH is 6.85.

Rick


  #3  
Old January 21st 05, 11:15 PM
Nikki Casali
external usenet poster
 
Posts: n/a
Default



Rick wrote:

"Nikki Casali" wrote in message
...

It's a month since I started using a permanent pH probe to control my
tank's CO2 injections. The probe being used is made by AquaMedic. At the
the first calibration I used buffer solutions 7 and 10. The instructions
say to use 7 and 9. I used the 10 instead to adjust the calibration
curve. The calibrated probe's readings matched the colorimeter's readings.

It's a month later. Time to recalibrate. I first checked to see how far
the probe had drifted. It was less than .1 according to the buffer
solutions. In this calibration routine I used 7 and 9.18. Calibration
done I then placed the probe into the tank water. It read 6.85. The last
reading before calibration was 7.25. I now checked with the colorimeter
which said the tank was 7.3. The probe is now off by -0.4!

I recalibrated with 7 and the original 10. Still reads 6.85!

It seems to me that I have somehow trashed the probe by recalibrating
it. Is there anything I can do to recover the probe? A temporary
solution I have, before I can get a replacement, is to calibrate the
probe to the known tank pH. That is, placing the probe in the tank water
and adjusting it until it reads around about 7.3.

Anyone have a similar experience?

Nikki



without using the calibration mode simply check the PH level of the buffer
you are using. You are then using a known solution. If the meter reads
correctly then your water PH is 6.85.


Thanks for the reply!

OK, I checked the 7.01 pH buffer solution with a colorimeter - the only
pH gadget I can rely upon. It gives 7.0 pH. I noticed that the pH probe,
which initially read 6.85 pH in the tank, drifted over 30 minutes to a
stable 7.05 pH. The colorimeter also concurs with a reading of 7.1 pH.
Everything is now in agreement. But a pH of 7.05 translates to 37 ppm of
CO2 with my water. Woops!

I reckon the buffer solutions pulled the probe's readings down
temporarily. Is this known to happen??


Nikki

  #4  
Old January 23rd 05, 07:30 PM
Rick
external usenet poster
 
Posts: n/a
Default



Thanks for the reply!

OK, I checked the 7.01 pH buffer solution with a colorimeter - the only
pH gadget I can rely upon. It gives 7.0 pH. I noticed that the pH probe,
which initially read 6.85 pH in the tank, drifted over 30 minutes to a
stable 7.05 pH. The colorimeter also concurs with a reading of 7.1 pH.
Everything is now in agreement. But a pH of 7.05 translates to 37 ppm of
CO2 with my water. Woops!

I reckon the buffer solutions pulled the probe's readings down
temporarily. Is this known to happen??


Nikki


I wouldn't be concerned with a 37 ppm of CO2. I keep mine around 30 all the
time. Great for the plants.

Rick


 




Thread Tools
Display Modes

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

vB code is On
Smilies are On
[IMG] code is On
HTML code is Off
Forum Jump

Similar Threads
Thread Thread Starter Forum Replies Last Post
Alternative pH calibration solutions? Any DIY recipes? John Reefs 7 August 4th 04 02:18 AM
CALIBRATION KIT 4.0ph, 7.0ph, 10.0ph -=Almazick=- Plants 10 October 8th 03 05:10 PM
Ice Probe Freezes Filter into Block of Ice logan Reefs 2 October 5th 03 04:26 PM
Grounding Probe Pszemol Reefs 54 September 1st 03 08:07 PM
Ice Probe + 7g Nano = Perfect Temp richard reynolds Reefs 0 July 25th 03 12:07 AM


All times are GMT +1. The time now is 03:08 PM.


Powered by vBulletin® Version 3.6.4
Copyright ©2000 - 2025, Jelsoft Enterprises Ltd.
Copyright ©2004-2025 FishKeepingBanter.com.
The comments are property of their posters.